taken from a cigar websiteI stumbled across somehow:

6.1 Checking your Hygrometer Calibration


Why should I check my hygrometer's calibration?
Even if you don't want to open its case and adjust your hygrometer, you would be well advised to check its calibration. These are not the precision instruments that some people think... Analog (dial-type) hygrometers are often shipped as much as 20% out of spec. Even the often-recommended Radio shack hygrometer is shipped with a factory tolerance of plus or minus 5 points of humidity through the range 40-80%. That means your unit could read anywhere from 65 to 75% while in a 70% environment and still be considered in spec! I've personally seen up to an 8 percentage point difference between units before calibration.

For a calibration source, we're going to use salt - NaCl - plain table salt... Why salt? The following is a brief explanation of the chemistry, originally posted to asc by Peter Shenkin of Columbia University. Don't worry too much about the technical details - the point is that plain salt will, when used properly will maintain an exact 75% humidity in a sealed environment.

A saturated solution at constant temperature & pressure has a fixed composition. [[ Explanation in terms of the Gibbs Phase Rule omitted. ]] Therefore, such a solution has a fixed vapor pressure. Thus, at constant temperature, no matter how much solid NaCl and how much water are present, the concentration of the NaCl in the water is fixed, just as long as both the solution and the solid phase are present. Therefore such a solution has a fixed vapor pressure.

Now, it just happens to turn out that the vapor pressure of a saturated solution of NaCl in H20 is about 75% of the vapor pressure of pure water at any temperature close to room temperature. This means that at equilibrium, if there were nothing present except this solution and a vapor phase in contact with it (no air), the pressure above the solution would be about 15 mm_Hg, in round numbers, using 20 mm_Hg as the vapor pressure of water near room temperature, again in round numbers. As mentioned earlier, the only gas providing the pressure above the solution would be water vapor. We'll see in a few minutes that the same thing happens even when air is introduced.

OK, class dismissed - you've got it now, right? :-) For more details about the science behind this calibration procedure, check out the excelent report available at: http://www.pip.dknet.dk/~pip1985/gemarkus.htm

Now to procedures. You'll need a zip-lock baggie or other see-through container, about a teaspoon of salt, and a small, shallow open container for that salt (I use the screw cap from a bottle of milk) Gee, I love this high-tech lab gear.... ;-)

Place a teaspoon or so of salt in its container and add a few drops of water to get it wet. You don't want to dissolve it, just get a good damp pile of salt in your bottle cap. This is important! Just a few drops! you want damp salt, NOT a solution. As the technical explanation above says, you want the salt present in a wet solid phase. Place it in the baggie, along with your hygrometer. CAUTION - DON'T SPILL ANY SALT ON YOUR INSTRUMENT! Be sure the Hygrometer isn't in its "min" or "max" mode, BTW.

Seal the baggie with some air trapped inside (so it's not tight against the hygrometer) and let it sit. Allow this to stabilize for at least 6 hours. (don't rush it!). After the internal "system" has stabilized, check your reading WITHOUT OPENING THE BAGGIE. It should be exactly 75%. If not, note the deviation - this is how much your hygrometer is out. Don't be surprised if your reading up to 5 points out - unfortunately, that's the factory tolerance of this instrument.

Note that several readers have had occasional problems with poorly sealing baggies. Use a small tupperware container, or double-bag your experimental setup will help avoid inaccuracies.


6.2 Radio Shack #63-855 Calibration

Ok! Feeling "scientific after testing the calibration of your hygrometer? Well, the next step for the ambitious among us is to actually callibrate our hygrometers. Many cybersmokers use the Radio Shack 63-855 Hygrometer, (also known as the "Airguide" Hygrometer). The following procedure was written specifically for that instrument. Please read these instructions very carefully before proceeding - If you don't feel comfortable tinkering with sensitive circuitry, DON'T DO IT!

Now that you know how far your instrument is out, you might want to adjust it. CAUTION! Mess-up here, and you could damage your hygrometer! There is no simple calibration knob to turn. To adjust this instrument, you'll be re-adjusting the bias current of the sensor, then comparing it to your calibration reference - the salt in the baggie.

Remove the battery door, the batteries, and finally the back of the unit. along the top edge of the PC board inside, you'll see two round, flattish metal disks (micro-potentiometers). don't touch the left one, you'll be working only with the one on the right. First, look at it closely. You'll see a tiny indent for a tiny screwdriver blade across the middle. Sketch its exact position carefully - this procedure might take several days, and you will forget the original orientation if you don't write it down!

You'll be turning this right potentiometer (pot) a tiny amount to change the humidity reading (1/8 turn might change your reading by 3 or 4%). DO NOT turn it more that 1/4 turn! Use a very small screwdriver (like a jeweler's screwdriver), and don't change it by much! If you were reading above 75% in the calibration test, turn this pot clockwise, again only a tiny amount... If your reading was low, turn it slightly counter-clockwise...

You can leave the back off to save some effort, and put the batteries back in now. Be sure the "min" and "max" functions are turned off, and place the hygrometer back in your baggie with the salt. Again, let it stabilize for 6 hours, and DON'T LET IT TOUCH THE WET SALT! Check your reading, and re-adjust if necessary. Be patient, you may have to try a few times to get it right. If you feel you're getting nowhere, set the pot back to exactly where it was when you started.

When you're satisfied with your setting (set it to +- 1% of 75%) remove the batteries, re-assemble the unit, and give it one final check. Once you're done, light up one of your best cigars - you'll deserve it!

Radio Shack #63-1013 Calibration

The Radio Shack model 63-1013 hygrometer can be calibrated similarly. The potentiometer (pot.) can be found on the upper left hand side of the circuit board once the back is removed. The pot. is small -- about half the diameter of a pencil eraser. The pot. has a small slot in it to allow for adjustments. Clockwise turns increase the humidy percent reading and counter clockwise turns decrease the reading.
6.3 Older Radio Shack units
contributed by Steven Banks

I would like to add some new information. I have one of the older Radio Shack hygrometers #63-844, but do you think that stopped me? You give me an opportunity to take something I know nothing about apart, I'm there! Plus, my Dad was a quality assurance engineer and he wouldn't have put up with that much of a error factor.

Well the salt portion of this whole procedure is sweet indeed - so simple, I felt like I was watching "Mr. Wizard" again. After doing this I found that my hygrometer was a whopping +8% off. Now... I'm not great with math, so I decided to correct the problem instead of just knowing how much it's off by. The problem was, which "pot" screws to adjust, since the instructions were for a #63-855. Well for those of you that have the same model as I do, there are two circuit boards in it, the top one is for the temperature and the bottom is for the humidity. The "pot" screw CLOSEST to the battery compartment is the one you need to adjust. And yes, as was mentioned in the first posting, it does take some fiddling!!! There were times I'd adjust the screw, and see no change. But I finally got that puppy within + -1%. Note that this hygrometer's reading will INCREASE as you turn the control CLOCKWISE - exactly the opposite of the 63-855 model...

Let it be known that I tinkered where no person in their right mind should, and this is no guarantee that it'll work for you (especially knowing Radio Shack stuff - yours could be laid out differently?) But if you're bold, go where no man has gone before and make it right!


6.4 The Airguide Digital Hygrometer

This is the same unit internally as the Radio Shack #63-855. See section 6.1 and 6.2.


6.5 The Bionaire Digital Hygrometer

This is a fine looking digital hygrometer available from several suppliers for about $17.95. It's similar in operation to the Radio Shack #63-855, buy without its memory features or internal adjustments. Its factory rated accuracy is +- 5%, but tests show its linearity (accuracy over a broad range) is considerably less. Since this unit can not be re-calibrated, I recommend testing its accuracy with the salt calibration test, to allow for differences in readings.


6.6 Analog Hygrometers

Surprisingly, both the most expensive and inexpensive classes of hygrometers are the dial-type (or analog) hygrometers. Many are sold in fancy brass or chrome housings, but their accuracy in general is not very impressive. Being a mechanical device, rough handling in shipping can often knock these gauges as far as 20% out from an accurate reading. The calibration of this type of instrument should always be checked.

Many of these gauges come with the simple instructions to "wrap the instrument in a damp towel for at least 20 minutes, then adjust it (either by set-screw or rotating the outer housing) to a reading of 95%." As this technique does not address the non-linearity of these gauges, I recommend checking them in the general RH range cigar-smokers need by using the above mentioned "salt-standard" calibration technique.

Don Meyer ([email protected]) has recommended a very inexpensive ($4.86) Sunbeam thermometer/hygrometer, available at Wal-Mart. On inspection in the store, he found that of the 10 or so on display, all read within 3%, whereas others (in the same store) displayed a broad range. Of course, consistency doesn't always mean accuracy, but it's a good sign. It should be noted that this unit is mounted in a rather large case. Users should consider re-mounting this unit for use in their humidors. The model # BTW, is 90113. (old #1803)

________________________________________________________________________
Top of pageBottom of pageLink to this message

Dr. Cubesis III (Newbieshroomer)
Posted on Friday, June 21, 2002 - 10:37 pm:Edit Post Quote Text Delete Post Print Post Move Post (Moderator/Admin Only)

Hey vitti...


How important would you say a hygrometer is?

When I first started, thought i absolutely had to have one.. But... Have been successful without one
with casings and cakes... So far so good..

So Vitti, what do you think?
How important is it?
Just curious my friend. You've been around a lot longer than I, and i'm interested in your answer...

Peace
Top of pageBottom of pageLink to this message

Vitti (Vitticeps)
Posted on Friday, June 21, 2002 - 10:48 pm:Edit Post Quote Text Delete Post Print Post Move Post (Moderator/Admin Only)

Well, for the curious or the inconfident newbie they are great. But as you said, once you get a feel for things you don't really need one so much. Either way they're nice to have. But having one that's giving bad readings is not good. :)
Top of pageBottom of pageLink to this message

Imok Urok2 (Imok)
Posted on Friday, June 21, 2002 - 11:21 pm:Edit Post Quote Text Delete Post Print Post Move Post (Moderator/Admin Only)

.